Default Fused Test Leads

We have an ongoing debate in our company... The ontario Electrical Safety Authority (ESA) recommends using fused test leads for multimeters. A Fluke representative at a recent electrical safety training seminar recommended against using fused test leads. Who is right and why?
Reply With Quote
  #2  
Old 10-12-2006, 01:08 PM
Glenn D Glenn D is offline
Senior Member
 
Join Date: Oct 2006
Location: South Florida
Posts: 488
Default Re: Fused Test Leads

I'd really like to understand why they would want fused test leads. Personal protection? Wouldnt a fuse add to the total resistance of the lead?. I have heard of fused jumpers... thats a good Idea.. On most meters the ampere scale is fuse protected. what other scale needs fuse protection?

Last edited by Glenn D : 10-12-2006 at 01:09 PM. Reason: mis statement
Reply With Quote
  #3  
Old 11-12-2006, 08:44 AM
MAZ's Avatar
MAZ MAZ is offline
Member
 
Join Date: Apr 2006
Posts: 63
Exclamation Re: Fused Test Leads

The further away the protection ,the less effective the protection is.Putting hi-energy fuses at the business end of the measuring circuit,YOUR HANDS!,gives you extra protection.What if your meter is dirty or there is internal contamination as encountered in dirty industrial environments?The arc fault can blow before the meters internal fuse.Yes,there is resistance in the fuse but that is what the REL function on ohms is for.To null out test lead resistance.Consider fused test leads as back up for the meters fuse.

I have no idea why Fluke would not recommend using fused test leads.
A meter can be a bomb when used in harsh environments and any extra protection is well worth it.I think fused test leads with proper ratings are a fantastic safety idea for test equipment users.
Reply With Quote
  #4  
Old 12-01-2006, 05:10 PM
76nemo's Avatar
76nemo 76nemo is offline
Forum Dummy
 
Join Date: Aug 2006
Location: Stuck in upstate NY.
Posts: 1,719
Default Re: Fused Test Leads

Just simply for my own curiosity, I would really get to the bottom of this. The more I ask, the more I learn. I want to hear from Fluke on this. Why would you possibly stray away from fused test leads? This to me isn't a head scratcher, I must be missing something as well. I totally agree with the post by MAZ. Fluke, do you have any kind response to this? And further more, Fluke does not offer fused test leads in their line of product, any particular reason?
Thanks guys,

~76nemo

Last edited by 76nemo : 08-09-2008 at 11:27 PM.
Reply With Quote
  #5  
Old 12-03-2006, 02:04 PM
MAZ's Avatar
MAZ MAZ is offline
Member
 
Join Date: Apr 2006
Posts: 63
Thumbs up Re: Fused Test Leads

Thanks for the backup 76 Nemo.I too would like a response from Fluke.
Anyone interested,the fused test leads I use are Silvertronic.Transcat.com has them.I think John Fluke would have liked the idea himself.
Reply With Quote
  #6  
Old 12-04-2006, 08:57 PM
76nemo's Avatar
76nemo 76nemo is offline
Forum Dummy
 
Join Date: Aug 2006
Location: Stuck in upstate NY.
Posts: 1,719
Default Re: Fused Test Leads

Still no response from this, except from MAZ. And as for the post about adding resistance to the circuit, again, you can't disagree with MAZ. The added resistance is way to small to cause any distraction in most circuit analysis. Fluke, we're listening. If you guys can send us some info. that we can't imagine, it would greatly be appreciated. I am one of the people who really enjoys a good debate.
Fused test leads,..... why not?

~76nemo
Reply With Quote
  #7  
Old 12-06-2006, 12:31 PM
DuaneS's Avatar
DuaneS DuaneS is offline
Moderator
 
Join Date: Sep 2005
Posts: 27
Default Re: Fused Test Leads

Quote:
Originally Posted by ASCO Phil
We have an ongoing debate in our company... The ontario Electrical Safety Authority (ESA) recommends using fused test leads for multimeters. A Fluke representative at a recent electrical safety training seminar recommended against using fused test leads. Who is right and why?


Here is our opinion on fused test leads:

It is Fluke's opinion that in applications using Fluke name brand instruments, fused test probes provide little if any added protection. Fluke's primary concern is that fused test probes may result in serious shock hazard do to a false voltage reading. A more effective solution is a meter properly designed to handle overload conditions. For the Amp and milliamp functions, Fluke meters incorporate a specifically designed and tested fuse that is effective to 1000 Volts as the full breaking capacity of the system which limits the energy in a manner that is as effective as a fuse located within a test probe.



It is additionally Fluke's position that a fused test probe is potentially more hazardous when used with any DMM for the following reason. It is estimated that over 90% of most electrician measurements use the AC VOLTS functions of a meter. Current measures are generally less than 10%, more probably in the few percentage points. If the fuse in the test probe is damaged and not noticed or properly corrected, then false voltage measurements will occur which potentially create a shock hazard for the user.



For Fluke meters, a damaged fuse within the amps function of the meter does not affect voltage measurements, so the potential of shock hazard due to false readings is eliminated. Note that for the other meter functions such as ohms, Fluke meters are designed to prevent hazard for overload conditions equivalent the maximum AC volts input as stated on the meter. These functions limit the energy by impedance. This opinion on fused test leads is limited to, an only applies to their use when used with Fluke brand test instruments

Duane
Reply With Quote
  #8  
Old 12-06-2006, 04:38 PM
76nemo's Avatar
76nemo 76nemo is offline
Forum Dummy
 
Join Date: Aug 2006
Location: Stuck in upstate NY.
Posts: 1,719
Default Re: Fused Test Leads

Question answered. What Maz and I did not include in our question was taking into consideration that a well experienced electrician was using the fused leads, not anyone who just pulled the meter out of the box, read it's safety limitations and tried to troubleshoot any ol' problem without the knowledge of proper procedures. Again, I love a good debate, and you got me on this is one,.....SAFETY FIRST TO EVERYONE!!!!!!!!
Who am I to question Fluke? I knew there had to be some kind of reason, and you nailed me with it. That's what you get for not thinking out of the box.
Thanks Fluke,

~76nemo
Reply With Quote
  #9  
Old 12-07-2006, 07:51 PM
MAZ's Avatar
MAZ MAZ is offline
Member
 
Join Date: Apr 2006
Posts: 63
Exclamation Re: Fused Test Leads

I agree with Nemo,on EXPERIENCE and Dwane on a possibly open fuse.I assumed that multimeter users go by one of the best safety rules I learned the hard way many years ago,TEST YOUR METER ON A KNOWN LIVE CIRCUIT BEFORE MEASURING THE LIVE CIRCUIT YOU ARE TESTING.This checks that your meter,leads,and FUSE in the test leads mentioned are working.
How many guys out there went nuts or even got zapped because of a flakey open test lead?Your leads get bent, twisted, and pulled every time you use them.
The newer leads have very fine multistrand wires that can easily be broken inside the leads insulation where you can't see the damage.
What I do is plug the meter leads into the meter, put the meter in continuity beep test,short the probe tips in one hand,and shake the meter with the leads hanging and bouncing.The beeper in my Fluke 27 will beep off and on sounding the flakey connection.
Reply With Quote
  #10  
Old 12-08-2006, 08:53 AM
DuaneS's Avatar
DuaneS DuaneS is offline
Moderator
 
Join Date: Sep 2005
Posts: 27
Default Re: Fused Test Leads

I agree with Maz. The best way to insure safety is to use what is referred to as the 3 point test method (test known live, test circuit under test, test known live again). This insures that your equipment is good and your meter is on the right function and range and your leads are in the right jacks ahead of the circuit test. This process eliminates the most common measurement mistakes and also verifies your meter and leads are functional.

Duane
Reply With Quote
Reply


Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

vB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Forum Jump



All times are GMT -7. The time now is 06:27 PM.


Powered by: vBulletin Version 3.0.8
Copyright ©2000 - 2010, Jelsoft Enterprises Ltd.